Good day Masters,

I just want to ask about Jedi HV tank of GE BS8.

Is it safe to open the tank and refill an hv oil?

Thank you for the reply




上一篇:飞利浦东软CT球管型号
下一篇:西门子somatom spirit CT 机故障维修
It would be better to rely on ourselves than on others.
回复

使用道具 举报

 楼主| 发表于 2018-12-3 21:00:24 | 显示全部楼层
The Tank contains a bellows on the bottom, this bellows is required to compensate for oil heating due to Tank usage. Upon opening the Tank, the bellows collapses. If, on refilling the Tank, when heated due to usage, oil will pour out onto the Gantry.
You can attempt to reset the bellows by various means, but without knowing the exact adjustment set at the factory, you run the risk of spewing oil due to heating.
Repair of the Tank requires the possibility of having to replace 400 diodes, two specialized resistors, two Bleeder resistors and 8 Capacitors rated 20KVDC, as well as 4 special High frequency secondary coils, 4 Primary coils, specialized ceramic core pieces and the KV Measure board which cannot be replaced without opening the Tank.

I have repaired several hundred of these, including ones where the customer opened the Tanks and claimed there was no oil, because the bellows collapsed and he could not see any oil. That particular client in the Philippines actually opened the Tanks twice, paying no attention to the advice I provided.
I can repair any Jedi Tank, but what is the problem you are experiencing indicating you need to go inside the Tank?
For instance, KV Unbalance implies you have likely lost a 50 diode bank, a coil or a resistor or a capacitor.

Let me know if I can help.

call me : (0)13926236686
It would be better to rely on ourselves than on others.